Page 1 of 1
pH meter calibration
Posted: 29 Oct 2026, 06:21
by Marama T
Does anyone else struggle with calibration of pH meters? I find that no matter how much I try to calibrate them, I won't get consistent readings (i.e. several meters sitting in the same solution at the same time will give different values).
Re: pH meter calibration
Posted: 29 Oct 2026, 08:44
by PeterK
This is generally due to a number of reasons.
Firstly, the pH probes schools use are quite poor quality. Better quality probes are quite expensive, and with student use, not really practical. Secondly, in analytical labs, probes a calibrated at least daily against known certified buffers. These solutions don't last long (maybe a month in the fridge) so not really viable for schools, especially when for day-to-day use, we can get away with just using pH test strips. There is also often no maintenance done on the probes (quite often the probe is enclosed so it won't let you refresh the electrode solution).
Lastly there is sampling technique, pH probes are designed to be used in moving water. If you are reading a pH place the sample on a stirrer or at least stir the probe. A probe you are using contaminates the sample, so it's best to use fresh sample for each test. If you wanted to test several probes divide a well-mixed sample into separate samples for each probe so that they are not interfering with each other.
There is a lot more detail than that, but that is the basics.
Re: pH meter calibration
Posted: 29 Oct 2026, 08:56
by bigmack
It is a reality of Digital pH meters and it's useful for students to see this so they can understand that the number they read is not a guarantee but a close approximation .
Having said that , we use Pasco pH probes and I have to say they are the best and most repeatable of any I have used .
I use the two point calibration with them . pH4 and pH10 .
I'm probably preaching to the converted , but pH meters don't give an instant reading . It takes time for the Ions to travel through the glass tip the . So any time you take a reading , you must stir it around in the solution for about 20-30 seconds before the reading stabilizes.
New probes may only take 10-15 seconds to stabilize whereas probes that are older than 2-3 years can take almost a minute .
This is important when calibrating them . If you just dip it into the pH buffer and click the calibrate button , it will be way off .
pH meters also don't like going from a high pH to a low pH and visa vera . They may take a minute or two to stabilize and it is important to rinse (give it a good stir ) in a beaker of Tap water between one reading and the next .
Never use distilled water around pH probes . It sucks the ions out of the internal solution. Always use tap water .
The Pasco probes I use , when I calibrate them , and test them later , I might get a range of readings from say 3.98 -4.09
Edit : posted with Peter
Re: pH meter calibration
Posted: 29 Oct 2026, 09:00
by Marama T
Never use distilled water??????? OMG maybe that's my problem - I've been rinsing them between readings in a beaker of distilled water. Dagnabbit. Thanks bigmack. I did know about the reading stabilization though.
Re: pH meter calibration
Posted: 29 Oct 2026, 09:27
by bigmack
Curious what probes you have Marama ?
Re: pH meter calibration
Posted: 29 Oct 2026, 09:39
by PeterK
Rinsing isn't too bad, as long as you use a wash bottle. It is more letting it have longer contact like storage or reading it as a sample or even resting it in distilled water. It's really depending on how you're using it. If you want precision, you might rinse with distilled. If you don't need it that precise rinse with tap. The trade-off for precision is more maintenance and generally at school precision isn't needed so tap water is best.
The Pasco pH probes are decent for their price. They are not cheap though. We had two. They went to camp with the biology excursion. Only one came back working. :(
Re: pH meter calibration
Posted: 29 Oct 2026, 11:11
by bigmack
PeterK wrote: 29 Oct 2026, 09:39
The Pasco pH probes are decent for their price. They are not cheap though. We had two. They went to camp with the biology excursion. Only one came back working. :(
I feel your pain Peter .
We had two destroyed in Yr11 chem .
Despite me reminding the teacher to tell the students to unscrew the storage bottle before removing the probe , they just ripped it out and the vacuum created punctures a hole in the thin glass tip .Bye bye $600

Re: pH meter calibration
Posted: 29 Oct 2026, 17:53
by Pm706Narrogin
*Sigh* This is why we can't have nice things
I sometimes wonder if some of these items were $600 personal items if teachers would be as lax with ensuring they're used properly. But I think I know the answer to that

Re: pH meter calibration
Posted: 30 Oct 2026, 09:09
by RosalieL
I can't remember which one is which, so this is potentially not very helpful, but there are 2 types of pH meters and one can be repaired and the other cannot. I think it was one of the sales reps from Ciderhouse Tech at a Labbies day I went to a few years ago who was talking about it so you could ask them. I think it is the ones with the two sensor things at the bottom (like the Hanna ones) are not repairable, but the ones with the bulb (like the pasco ones) can be repaired by getting the new bulb.
Re: pH meter calibration
Posted: 02 Nov 2026, 06:36
by Marama T
bigmack wrote: 29 Oct 2026, 09:27
Curious what probes you have Marama ?
Five from Hanna that are pretty good. One from Haines that is ok. Made by a company called Analytical Instruments. It was $300 and when it turned up I googled the company because it looks cheap and badly made. They're based in Sri Lanka. Glad I only ordered one, honestly. I can recommend NOT buying these.
Re: pH meter calibration
Posted: 02 Nov 2026, 08:50
by bigmack
How old are your Hanna ones Marama ?
we had one of those at my old school when I started there over 12 years ago . even then , it was very sluggish .The new version of them has software calibration instead of the presets that you adjust with a screwdriver . But heck they are expensive .
Re: pH meter calibration
Posted: 02 Nov 2026, 09:20
by Marama T
I'm not sure how old they are. They're just the little pHep pocket waterproof testers. I started here in 2016 so I would hazard a guess at four years old. I calibrate them by pressing a button and putting them in a pH 7 buffer solution. We bought ten originally and five still work.
Re: pH meter calibration
Posted: 04 Nov 2026, 22:29
by bigmack
My bad Marama , I looked at your pic and thought it was the old Hanna meter we had . I guess that is the Haines (Hanna clone) you were referring to as “cheap and badly made .”
IMG_5171.jpeg
This is reminding me of the gas syringe look-a-like trap I fell down a month or so ago .

Re: pH meter calibration
Posted: 05 Nov 2026, 06:42
by Marama T
Oh WOW! My one is a rip-off version! Should be illegal.Crap.
Re: pH meter calibration
Posted: 10 Nov 2026, 11:27
by bigmack
Got ours out for a titration prac and checked calibration . Last checked in August . 3 out of 4 were still within 0.02 when I checked them with the Pasco buffer solution I had made up back then .
Just out of interest I made up some fresh buffer with tablets from Perth Scientific .
This is the interesting thing . The pH 10 and 7 read exactly the same as the Pasco buffer , but the pH 4 read 4.07- 4.10
I made up a fresh Pasco pH 4 just incase it had gone off , but it was exactly the same as the 3 month old one .
So it has raised the question about how accurate are the buffer tablets we buy ? Has anyone else found them to be different .
IMG_5200.jpeg